Repeatability versus intermediate precision — what changed since

DO
dr_okonkwoTL4 Moderator15 Jun 2026#1

Posting this under the heading it deserves: Repeatability versus intermediate precision — what changed since Everything below is what sits behind that.

Working through the identity arithmetic and I would like it checked.

tirzepatide has a monoisotopic mass close to 4813.5 Da. On an electrospray instrument I would expect to see the multiply charged series rather than the intact singly charged ion, so for the doubly charged species I calculate (4813.5 + 2 x 1.00728) / 2, and for the triply charged the analogous expression.

The observed values in the report sit within a few ppm of those. My question is what that actually establishes, because I have seen people treat a mass match as a purity result and I do not think it is one.

1 like 1mo
BR
buffer_reviewTL3Regular30 Jun 2026#2

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

5 likes 28d
AN
a.norgaardTL2 Moderator10 Jul 2026#3

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

20 likes 18d
TI
trough_indexTL3Regular19 Jul 2026#4
a.norgaard, post #3: System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5). Go to post

Coming back to post #3, because the follow-up matters more than the original answer.

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

0 likes in reply to #3 9d
FN
f.novakTL2 Moderator28 Jul 2026#5

post #4 is right about the mechanism and I think understates the practical bit.

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

0 likes 5h

Suggested topics

TopicParticipantsRepliesViewsActivity
Specificity, linearity, accuracy, precision, range, robustness — with real criteria
Specificity, linearity, accuracy, precision, range, robustness — with real criteria Writing it up because I had to work it out twice and would rather nobody else did. I would like to understand what this…
FNISMLAWM+44 48 12k 1d
Why one laboratory's 98.4% and another's 96.1% may not contradict
Why one laboratory's 98.4% and another's 96.1% may not contradict I have a specific reason for asking rather than idle curiosity, and the context is below. A documentation question rather than an analytical…
LCMEHMLPFC+28 37 457 7mo
Why one laboratory's 98.4% and another's 96.1% may not contradict — one year on
Asking directly, because I could not find a straight answer: Why one laboratory's 98.4% and another's 96.1% may not contradict — one year on Working through the identity arithmetic and I would like it…
MSFKJBMRCR+10 14 13k 5mo
Coming back to: When to suspect the method rather than the sample
When to suspect the method rather than the sample I have a specific reason for asking rather than idle curiosity, and the context is below. Posting the method first, because I know what the first three…
OVGBFRSLRM+116 135 42k 19mo
Specificity, linearity, accuracy, precision, range, robustness — with real criteria — a second dataset
On the subject in the title: Specificity, linearity, accuracy, precision, range, robustness — with real criteria — a second dataset Working notes rather than a conclusion. Working through the identity…
MSHFDBTDLG+53 58 7.5k 6mo

Related topics — sharing the tags recovery, precision & repeatability, linearity

TopicParticipantsRepliesViewsActivity
[2026 update] Why 99.2% and 97.8% on the same lot can both be correct
Why 99.2% and 97.8% on the same lot can both be correct — that is the question, and I have not found it answered plainly anywhere I have looked. I would like to understand what this number means before I…
BBWSZDSTD+1 5 719 2d
Forced degradation studies: designing one that is informative
Posting this under the heading it deserves: Forced degradation studies: designing one that is informative Everything below is what sits behind that. Working through the identity arithmetic and I would like it…
MPTIEMHKSL+100 106 35k 2d
Specificity, linearity, accuracy, precision, range, robustness — with real criteria
Specificity, linearity, accuracy, precision, range, robustness — with real criteria Writing it up because I had to work it out twice and would rather nobody else did. I would like to understand what this…
FNISMLAWM+44 48 12k 1d
Injection volume, overload, and peak distortion — one year on
Injection volume, overload, and peak distortion — one year on Writing it up because I had to work it out twice and would rather nobody else did. A documentation question rather than an analytical one. I have…
DTCNFEGFZV+4 8 1.2k 4mo
Why one laboratory's 98.4% and another's 96.1% may not contradict
Why one laboratory's 98.4% and another's 96.1% may not contradict I have a specific reason for asking rather than idle curiosity, and the context is below. A documentation question rather than an analytical…
LCMEHMLPFC+28 37 457 7mo