Analytics · Mass spectrometry · continued

Distinguishing a deletion sequence by mass alone — the long version posts 31–45

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1.

IS
isotonic_sheetTL3Regular4 Jul 2025#31

I read post #29 twice before replying, because I had assumed the opposite.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

0 likes 13mo
PK
p.krastevTL2 Moderator8 Jul 2025#32

This follows post #29 rather than contradicting it.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

25 likes 13mo
RV
r.venkatesanTL3Wiki editor13 Jul 2025#33
e.silva, post #6: Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples. Go to post

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

12 likes in reply to #6 13mo
FD
f.danquahTL2 Moderator17 Jul 2025 · edited#34

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

4 likes 12mo
TK
t.kulkarniTL321 Jul 2025#35
AV
a.vermeulenTL2 Moderator26 Jul 2025#36

Picking up post #33: that is the part I would want checked first.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes 12mo
RJ
r.jhannsdttirTL3Regular30 Jul 2025#37
p.frisk, post #3: post #2 is right about the mechanism and I think understates the practical bit. Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very… Go to post

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

17 likes in reply to #3 12mo
NK
ni.kravchenkoTL2 Moderator3 Aug 2025#38
isotonic_sheet, post #31: I read post #29 twice before replying, because I had assumed the opposite. Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

7 likes in reply to #31 12mo
CL
coldchain_liuTL3Regular8 Aug 2025#39

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

1 like 12mo
MN
m.nascimentoTL2 Moderator12 Aug 2025#40

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

0 likes 12mo
CP
citation_peakTL316 Aug 2025#41
AC
a.cabreraTL2 Moderator20 Aug 2025 · edited#42
methods_margin, post #15: Two things before anyone answers the substance. First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound. Go to post

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

3 likes in reply to #15 11mo
KR
k.redgraveTL2Member24 Aug 2025#43
n.krastev, post #29: post #28 answers the question as asked. The question underneath it is different. Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

post #42 answers the question as asked. The question underneath it is different.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

17 likes in reply to #29 11mo
SI
s.ivaturiTL2 Moderator29 Aug 2025#44

On post #40 — agreed on the reasoning, with one qualification.

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

32 likes 11mo
R
RidgewayTL3Regular2 Sep 2025#45

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes 11mo

Suggested topics

TopicParticipantsRepliesViewsActivity
Second pass at: What a 2 ppm match establishes, and what it does not
Second pass at: What a 2 ppm match establishes, and what it does not — setting out what I have, and where I think it stops being reliable. Working through the identity arithmetic and I would like it checked.…
BSNKSTMGMM+44 49 8.1k 1d
Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer I have a specific reason for asking rather than idle curiosity, and the context is below. Posting the method first, because I know what the first three…
DAAPINOSS+75 82 4k 17mo
MALDI versus ESI for a quick identity check
On the subject in the title: MALDI versus ESI for a quick identity check Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. tirzepatide has a…
RMTVO 2 2k 11mo
Distinguishing a deletion sequence by mass alone
Distinguishing a deletion sequence by mass alone Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it checked.…
RMLSMHMINM+121 129 25k 17mo
About the Mass spectrometry category
LC-MS and high-resolution MS, charge states, mass accuracy in ppm, and what identity confirmation does and does not establish. This post is a community wiki: any member at trust level 3 or above can edit it,…
DOOBLDBDRV+5 9 19k 6h

Related topics — sharing the tags high-resolution MS, mass error (ppm), charge states

TopicParticipantsRepliesViewsActivity
About the Mass spectrometry category
LC-MS and high-resolution MS, charge states, mass accuracy in ppm, and what identity confirmation does and does not establish. This post is a community wiki: any member at trust level 3 or above can edit it,…
DOOBLDBDRV+5 9 19k 6h
Mass error in ppm: the arithmetic and the caveat — a second dataset
On the subject in the title: Mass error in ppm: the arithmetic and the caveat — a second dataset Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked.…
HKGVDWONW+20 24 1.7k 1d
[2026 update] Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
GTAPFP 2 1k 8mo
Isotope patterns and why the monoisotopic peak is not the tallest — does this still hold?
Isotope patterns and why the monoisotopic peak is not the tallest — does this still hold? — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity…
PODBTD 2 4.6k 7h
Coming back to: Deconvolution of a multiply charged envelope
On the subject in the title: Deconvolution of a multiply charged envelope Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. tirzepatide has a…
VRCVEL 2 12k 22mo