Distinguishing a deletion sequence by mass alone — the long version

TF
taper_fileTL3Regular13 Jan 2025#1

On the subject in the title: Distinguishing a deletion sequence by mass alone — the long version Working notes rather than a conclusion.

I would like to understand what this number means before I repeat it anywhere.

A VendorInvestigate report on a semaglutide lot gives 99% purity. The supplier certificate for the same lot states 98.6%. Both documents name a reversed-phase method; neither states the same gradient.

My question is not "who is right". It is: given that those two figures were produced by different methods, what is the largest difference I should expect from method alone, and at what point does a gap stop being explainable that way?

11 likes 18mo
VK
v.klausenTL3Regular25 Jan 2025#2

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

16 likes 18mo
PF
p.friskTL2 Moderator2 Feb 2025 · edited#3

post #2 is right about the mechanism and I think understates the practical bit.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

31 likes 18mo
ID
integrator_draftTL3Regular10 Feb 2025#4
p.frisk, post #3: post #2 is right about the mechanism and I think understates the practical bit. Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very… Go to post

Worth separating two things that the opening post runs together.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

0 likes in reply to #3 18mo
LF
l.ferreiraTL2 Moderator17 Feb 2025#5

Picking up post #2: that is the part I would want checked first.

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

10 likes 17mo
ES
e.silvaTL2 Moderator24 Feb 2025#6

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

22 likes 17mo
SS
s.solbergTL2 Moderator3 Mar 2025#7
p.frisk, post #3: post #2 is right about the mechanism and I think understates the practical bit. Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very… Go to post

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes in reply to #3 17mo
HK
h.koodziejTL2Member9 Mar 2025#8
l.ferreira, post #5: Picking up post #2: that is the part I would want checked first. For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use. Go to post

On post #4 — agreed on the reasoning, with one qualification.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

1 like in reply to #5 17mo
KB
ka.batistaTL2 Moderator15 Mar 2025#9

This follows post #6 rather than contradicting it.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

15 likes 16mo
SF
sterile_fileTL3Regular21 Mar 2025 · edited#10

I read post #8 twice before replying, because I had assumed the opposite.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

30 likes 16mo
AT
a.thorneTL2Wiki editor27 Mar 2025#11

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes 16mo
HF
h.friskTL2 Moderator1 Apr 2025#12

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

21 likes 16mo
CR
crossover_reviewTL3Regular7 Apr 2025#13
integrator_draft, post #4: Worth separating two things that the opening post runs together. Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity. Go to post

I read post #11 twice before replying, because I had assumed the opposite.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

5 likes in reply to #4 16mo
JS
j.sandvikTL2 Moderator12 Apr 2025#14

This follows post #11 rather than contradicting it.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 16mo
MM
methods_marginTL3Regular18 Apr 2025#15

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

29 likes 15mo
EB
e.bakkenTL2 Moderator23 Apr 2025#16

post #15 answers the question as asked. The question underneath it is different.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

15 likes 15mo
I
IRenaudinTL2Member28 Apr 2025 · edited#17
s.solberg, post #7: Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Coming back to post #15, because the follow-up matters more than the original answer.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

3 likes in reply to #7 15mo
SC
s.chowdhuryTL3Regular3 May 2025#18
integrator_draft, post #4: Worth separating two things that the opening post runs together. Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity. Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

0 likes in reply to #4 15mo
AK
a.kwiatkowskiTL2Member8 May 2025#19

Worth separating two things that post #15 runs together.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 15mo
KO
k.ogunleyeTL2 Moderator13 May 2025#20
ka.batista, post #9: This follows post #6 rather than contradicting it. Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes in reply to #9 15mo
GR
g.radichTL2 Moderator18 May 2025#21

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes 14mo
MM
maintenance_modeTL3Regular23 May 2025 · edited#22

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

4 likes 14mo
YA
y.adeyemiTL2 Moderator28 May 2025#23
l.ferreira, post #5: Picking up post #2: that is the part I would want checked first. For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use. Go to post

Picking up post #20: that is the part I would want checked first.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

13 likes in reply to #5 14mo
RV
r.venkatesanTL3Wiki editor2 Jun 2025#24
sterile_file, post #10: I read post #8 twice before replying, because I had assumed the opposite. What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

Coming back to post #22, because the follow-up matters more than the original answer.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

27 likes in reply to #10 14mo
HB
h.brandtTL2 Moderator6 Jun 2025#25

post #24 is right about the mechanism and I think understates the practical bit.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

2 likes 14mo
IS
isotonic_sheetTL3Regular11 Jun 2025#26

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

8 likes 14mo
PT
p.trevinoTL2 Moderator16 Jun 2025#27

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

19 likes 13mo
R
RodriguesTL3Regular20 Jun 2025#28
ka.batista, post #9: This follows post #6 rather than contradicting it. Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

I read post #26 twice before replying, because I had assumed the opposite.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes in reply to #9 13mo
NK
n.krastevTL2 Moderator25 Jun 2025 · edited#29

post #28 answers the question as asked. The question underneath it is different.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

4 likes 13mo
BD
baseline_driftTL2Analytical chemist29 Jun 2025#30

On post #26 — agreed on the reasoning, with one qualification.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

12 likes 13mo