Mass error in ppm: the arithmetic and the caveat — a second dataset

Solved
Solved by ca.vermeulen in post #8
Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

Jump to the accepted answer →

HK
h.kimaniTL2 Moderator15 May 2026#1

On the subject in the title: Mass error in ppm: the arithmetic and the caveat — a second dataset Working notes rather than a conclusion.

Working through the identity arithmetic and I would like it checked.

retatrutide has a monoisotopic mass close to 4731.3 Da. On an electrospray instrument I would expect to see the multiply charged series rather than the intact singly charged ion, so for the doubly charged species I calculate (4731.3 + 2 x 1.00728) / 2, and for the triply charged the analogous expression.

The observed values in the report sit within a few ppm of those. My question is what that actually establishes, because I have seen people treat a mass match as a purity result and I do not think it is one.

1 like 2mo
GV
g.verhoevenTL2 Moderator21 May 2026#2

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

5 likes 2mo
DW
diluent_watchTL2Member26 May 2026#3

post #2 answers the question as asked. The question underneath it is different.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

13 likes 2mo
ON
o.nybergTL2 Moderator29 May 2026#4

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

28 likes 2mo
W
WoodhouseTL2Member2 Jun 2026#5

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

0 likes 2mo
ZV
z.vogelTL2 Moderator5 Jun 2026#6
h.kimani, post #1: On the subject in the title: Mass error in ppm: the arithmetic and the caveat — a second dataset Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. retatrutide has a monoisotopic mass close to 4731.3 Da. On an electrospray instrument I would expect to see the multiply charged… Go to post

I read post #4 twice before replying, because I had assumed the opposite.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

2 likes in reply to #1 2mo
GF
gradient_fileTL2Member9 Jun 2026 · edited#7

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

9 likes 2mo
CV
ca.vermeulenTL2 Moderator Solution12 Jun 2026#8

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

20 likes 2mo
RV
r.villalobosTL2 Moderator15 Jun 2026#9
ca.vermeulen, post #8: Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

Picking up post #6: that is the part I would want checked first.

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

0 likes in reply to #8 1mo
RD
r.danquahTL2 Moderator18 Jun 2026#10

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes 1mo
LP
l.piresTL2 Moderator21 Jun 2026#11
Woodhouse, post #5: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes in reply to #5 1mo
ME
m.ekstromTL2 Moderator23 Jun 2026#12
g.verhoeven, post #2: Resolution: "high resolution" commonly means Go to post

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

20 likes in reply to #2 1mo
BD
b.demirTL226 Jun 2026#13
HM
h.mensahTL2 Moderator29 Jun 2026#14

Picking up post #11: that is the part I would want checked first.

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

0 likes 29d
T
ThibodeauTL3Regular2 Jul 2026#15
Woodhouse, post #5: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Worth separating two things that post #11 runs together.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes in reply to #5 26d
FC
f.chowdhuryTL2 Moderator4 Jul 2026#16

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

27 likes 24d
ME
m.eriksenTL2 Moderator7 Jul 2026#17

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

8 likes 21d
HA
h.amankwahTL2 Moderator9 Jul 2026#18

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

2 likes 19d
RS
r.serranoTL2 Moderator12 Jul 2026 · edited#19

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

21 likes 16d
OB
owen.bradyTL4 Moderator14 Jul 2026#20
ca.vermeulen, post #8: Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

post #19 answers the question as asked. The question underneath it is different.

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

9 likes in reply to #8 14d
AM
a.mwangiTL2 Moderator17 Jul 2026#21
z.vogel, post #6: I read post #4 twice before replying, because I had assumed the opposite. What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

post #20 is right about the mechanism and I think understates the practical bit.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

26 likes in reply to #6 11d
B
BGiordanoTL2Member19 Jul 2026#22
l.pires, post #11: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Worth separating two things that post #18 runs together.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

0 likes in reply to #11 9d
BV
b.vestergaardTL2 Moderator22 Jul 2026 · edited#23

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

2 likes 6d
CD
c.delgadoTL2 Moderator24 Jul 2026#24

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

8 likes 4d
HB
h.bhattacharyaTL2 Moderator26 Jul 2026#25
r.danquah, post #10: Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples. Go to post

post #24 answers the question as asked. The question underneath it is different.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

19 likes in reply to #10 1d

Suggested topics

TopicParticipantsRepliesViewsActivity
High-resolution MS on aggregates and what it can see
Posting this under the heading it deserves: High-resolution MS on aggregates and what it can see Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked.…
CRGISCBPEC+117 123 6.5k 8mo
Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. retatrutide has a…
JPERBEBWBJ+124 146 15k 16mo
Second pass at: MALDI versus ESI for a quick identity check
Second pass at: MALDI versus ESI for a quick identity check Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it…
MALKO 2 18k 7mo
Charge states for a 4 kDa peptide, worked through — the long version
Charge states for a 4 kDa peptide, worked through — the long version Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like…
HEGTMI 2 293 3mo
[2026 update] Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
GTAPFP 2 1k 8mo

Related topics — sharing the tags counter-ion content, mass error (ppm), charge states

TopicParticipantsRepliesViewsActivity
Racemisation and why it is invisible to a mass measurement
Posting this under the heading it deserves: Racemisation and why it is invisible to a mass measurement Everything below is what sits behind that. Working through the identity arithmetic and I would like it…
NTKCTDOPA+120 127 41k 1mo
High-resolution MS on aggregates and what it can see
Posting this under the heading it deserves: High-resolution MS on aggregates and what it can see Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked.…
CRGISCBPEC+117 123 6.5k 8mo
Coming back to: Oxidation pathways for methionine and tryptophan
Oxidation pathways for methionine and tryptophan Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it checked.…
YRFHKRZSCP+1 5 36k 15mo