AK
a.kravchenkoTL2 Moderator12 Oct 2025#91
l.dziedzic, post #75: What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

I read post #89 twice before replying, because I had assumed the opposite.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

13 likes in reply to #75 10mo
B
BirkelandTL3Regular13 Oct 2025#92

This follows post #89 rather than contradicting it.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

4 likes 9mo
VR
v.rautioTL2 Moderator15 Oct 2025#93

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes 9mo
CT
cannula_traceTL3Regular16 Oct 2025#94
c.amankwah, post #16: Picking up post #13: that is the part I would want checked first. Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes in reply to #16 9mo
GA
g.amankwahTL217 Oct 2025#95
OF
outline_firstTL3Wiki editor18 Oct 2025#96

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

7 likes 9mo
CC
c.castellanosTL2 Moderator19 Oct 2025 · edited#97
j.fonseca, post #48: Coming back to post #46, because the follow-up matters more than the original answer. Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

1 like in reply to #48 9mo
NE
n.ekstromTL2Regular20 Oct 2025#98
a.kirchner, post #58: post #57 is right about the mechanism and I think understates the practical bit. Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of… Go to post

post #97 answers the question as asked. The question underneath it is different.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

0 likes in reply to #58 9mo
RC
r.coelhoTL2 Moderator21 Oct 2025#99

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

25 likes 9mo
EM
endpoint_marginTL2Member23 Oct 2025#100
Leiterman, post #40: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

12 likes in reply to #40 9mo
DV
d.vukovicTL2 Moderator24 Oct 2025#101

Worth separating two things that post #97 runs together.

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

23 likes 9mo
LG
lc_gradientTL3Analytical chemist25 Oct 2025#102

post #101 is right about the mechanism and I think understates the practical bit.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

10 likes 9mo
RM
r.mensaTL2 Moderator26 Oct 2025#103

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

3 likes 9mo
DB
dr_bhattacharyaTL3Physician27 Oct 2025#104
v.nascimento, post #22: I read post #20 twice before replying, because I had assumed the opposite. Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes in reply to #22 9mo
CB
c.boatengTL228 Oct 2025#105
MP
mira.patelTL4 Admin29 Oct 2025#106

post #105 answers the question as asked. The question underneath it is different.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

15 likes 9mo
RZ
r.zielinskiTL2 Moderator30 Oct 2025#107
a.kirchner, post #58: post #57 is right about the mechanism and I think understates the practical bit. Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of… Go to post

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

6 likes in reply to #58 9mo
RA
r.aldana_pharmdTL4Pharmacist1 Nov 2025 · edited#108
desiccant_notes, post #34: Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency. Go to post

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

1 like in reply to #34 9mo
BF
b.fonsecaTL2 Moderator2 Nov 2025 · edited#109
sleep_log, post #11: Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

11 likes in reply to #11 9mo
PM
p.marchettiTL2 Moderator3 Nov 2025#110

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

3 likes 9mo
BM
buffer_marginTL3Regular4 Nov 2025#111
g.ibarra, post #2: Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

32 likes in reply to #2 9mo
KH
k.haddadTL25 Nov 2025#112
EC
excursion_checkTL3Regular6 Nov 2025#113

post #112 is right about the mechanism and I think understates the practical bit.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

3 likes 9mo
AH
a.hartmannTL2 Moderator7 Nov 2025#114

Worth separating two things that post #110 runs together.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

11 likes 9mo
VK
v.krastevTL2 Moderator8 Nov 2025#115

Picking up post #112: that is the part I would want checked first.

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

0 likes 9mo
MA
m.achebeTL2 Moderator9 Nov 2025#116
Leiterman, post #40: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

1 like in reply to #40 9mo
P
PSkarbekTL3Regular11 Nov 2025#117

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

6 likes 9mo
AS
a.silvaTL2 Moderator12 Nov 2025#118

On post #114 — agreed on the reasoning, with one qualification.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

17 likes 8mo
VD
vial_deskTL3Regular13 Nov 2025#119
l.wikstrom, post #27: Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples. Go to post

This follows post #116 rather than contradicting it.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

17 likes in reply to #27 8mo
EM
e.mensaTL2 Moderator14 Nov 2025#120

I read post #118 twice before replying, because I had assumed the opposite.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

0 likes 8mo