High-resolution MS on aggregates and what it can see

CR
crossover_reviewTL3Regular31 May 2025#1

Posting this under the heading it deserves: High-resolution MS on aggregates and what it can see Everything below is what sits behind that.

Working through the identity arithmetic and I would like it checked.

tirzepatide has a monoisotopic mass close to 4813.5 Da. On an electrospray instrument I would expect to see the multiply charged series rather than the intact singly charged ion, so for the doubly charged species I calculate (4813.5 + 2 x 1.00728) / 2, and for the triply charged the analogous expression.

The observed values in the report sit within a few ppm of those. My question is what that actually establishes, because I have seen people treat a mass match as a purity result and I do not think it is one.

1 like 14mo
GI
g.ibarraTL2 Moderator4 Jun 2025#2

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

4 likes 14mo
SC
s.cardosoTL2 Moderator7 Jun 2025#3

This follows post #2 rather than contradicting it.

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

18 likes 14mo
BP
b.petrovTL2 Moderator10 Jun 2025 · edited#4

I read post #3 twice before replying, because I had assumed the opposite.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes 14mo
EC
e.coelhoTL2 Moderator12 Jun 2025#5

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

1 like 14mo
VK
v.klausenTL3Regular14 Jun 2025#6

On post #2 — agreed on the reasoning, with one qualification.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

8 likes 13mo
CH
ca.haddadTL2 Moderator16 Jun 2025#7
crossover_review, post #1: Posting this under the heading it deserves: High-resolution MS on aggregates and what it can see Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked. tirzepatide has a monoisotopic mass close to 4813.5 Da. On an electrospray instrument I would expect to see the multiply charged… Go to post

Picking up post #4: that is the part I would want checked first.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

25 likes in reply to #1 13mo
G
GDashwoodTL3Regular19 Jun 2025#8

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

0 likes 13mo
JM
j.mwangiTL4 Moderator21 Jun 2025#9

post #8 is right about the mechanism and I think understates the practical bit.

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

0 likes 13mo
DE
d.eriksenTL2 Moderator22 Jun 2025#10

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

0 likes 13mo
SL
sleep_logTL2Regular24 Jun 2025#11
ca.haddad, post #7: Picking up post #4: that is the part I would want checked first. Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

8 likes in reply to #7 13mo
LV
l.vukovicTL2 Moderator26 Jun 2025 · edited#12

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

2 likes 13mo
SC
s.chowdhuryTL3Regular28 Jun 2025#13

Worth separating two things that post #9 runs together.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

0 likes 13mo
AI
an.ibarraTL2 Moderator30 Jun 2025#14

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

27 likes 13mo
FN
formulary_notesTL3Regular2 Jul 2025#15
sleep_log, post #11: Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

13 likes in reply to #11 13mo
CA
c.amankwahTL2 Moderator3 Jul 2025#16

Picking up post #13: that is the part I would want checked first.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

4 likes 13mo
TH
TL4_HalvorsenTL4Leader · Journal club5 Jul 2025#17

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes 13mo
RE
r.ekstromTL2 Moderator7 Jul 2025#18
crossover_review, post #1: Posting this under the heading it deserves: High-resolution MS on aggregates and what it can see Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked. tirzepatide has a monoisotopic mass close to 4813.5 Da. On an electrospray instrument I would expect to see the multiply charged… Go to post

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

0 likes in reply to #1 13mo
EF
endo_fellow_rkTL3Endocrinology fellow8 Jul 2025 · edited#19
formulary_notes, post #15: Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing. The correction was fair and I had been repeating something I had not checked carefully enough. Go to post

I read post #17 twice before replying, because I had assumed the opposite.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

2 likes in reply to #15 13mo
YA
y.adebayoTL2 Moderator10 Jul 2025#20
l.vukovic, post #12: Resolution: "high resolution" commonly means Go to post

This follows post #17 rather than contradicting it.

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

0 likes in reply to #12 13mo
NR
n.rahimiTL2 Moderator12 Jul 2025#21
y.adebayo, post #20: This follows post #17 rather than contradicting it. I disagree with the reply above, and I think the disagreement is substantive rather than terminological. The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am… Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

5 likes in reply to #20 13mo
VN
v.nascimentoTL2 Moderator13 Jul 2025#22

I read post #20 twice before replying, because I had assumed the opposite.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

14 likes 13mo
IC
i.coelhoTL2 Moderator15 Jul 2025 · edited#23

post #22 is right about the mechanism and I think understates the practical bit.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes 12mo
PA
p.amankwahTL2 Moderator16 Jul 2025#24

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 12mo
K
KLindqvistTL4 Moderator18 Jul 2025#25
s.cardoso, post #3: This follows post #2 rather than contradicting it. For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use. Go to post

Picking up post #22: that is the part I would want checked first.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

8 likes in reply to #3 12mo
CT
c.tullochTL2 Moderator20 Jul 2025#26
j.mwangi, post #9: post #8 is right about the mechanism and I think understates the practical bit. I disagree with the reply above, and I think the disagreement is substantive rather than terminological. The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this,… Go to post

Coming back to post #24, because the follow-up matters more than the original answer.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

20 likes in reply to #9 12mo
LW
l.wikstromTL2 Moderator21 Jul 2025#27

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes 12mo
FK
f.kimaniTL2 Moderator23 Jul 2025#28

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

2 likes 12mo
BD
baseline_driftTL2Analytical chemist24 Jul 2025#29

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

13 likes 12mo
IA
id.almeidaTL2 Moderator26 Jul 2025#30
crossover_review, post #1: Posting this under the heading it deserves: High-resolution MS on aggregates and what it can see Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked. tirzepatide has a monoisotopic mass close to 4813.5 Da. On an electrospray instrument I would expect to see the multiply charged… Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

27 likes in reply to #1 12mo