Coming back to: When to suspect the method rather than the sample posts 121–136

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1 · go to the accepted answer.

RS
r.serranoTL2 Moderator23 Dec 2024#121
Ostrowski, post #87: Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance. Go to post

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

1 like in reply to #87 19mo
MH
ms_hollowayTL4Mass spectrometrist23 Dec 2024 · edited#122

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

7 likes 19mo
KD
k.dahlbergTL2 Moderator24 Dec 2024#123

post #122 answers the question as asked. The question underneath it is different.

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

24 likes 19mo
OB
owen.bradyTL4 Moderator24 Dec 2024#124

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

0 likes 19mo
BO
b.oseiTL2 Moderator24 Dec 2024#125
p.krastev, post #103: This follows post #100 rather than contradicting it. Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance. Go to post

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

0 likes in reply to #103 19mo
AR
a.reyesTL4 Admin24 Dec 2024#126
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

I read post #124 twice before replying, because I had assumed the opposite.

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

4 likes 19mo
BD
b.demirTL2 Moderator24 Dec 2024#127

Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99.

18 likes 19mo
SB
s.bruunTL2 Moderator24 Dec 2024#128

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

0 likes 19mo
CD
c.delgadoTL2 Moderator25 Dec 2024#129

Picking up post #126: that is the part I would want checked first.

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

0 likes 19mo
AL
a.lindholmTL2 Moderator25 Dec 2024#130
j.vandermolen, post #95: Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong. Go to post

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

1 like in reply to #95 19mo
TT
taper_tableTL3Regular25 Dec 2024#131

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

0 likes 19mo
TV
t.verhoevenTL2 Moderator25 Dec 2024#132

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

32 likes 19mo
EC
excursion_checkTL3Regular25 Dec 2024 · edited#133

Coming back to post #131, because the follow-up matters more than the original answer.

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

11 likes 19mo
SV
s.vanheckeTL2 Moderator25 Dec 2024#134
Knowlton, post #33: Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Picking up post #131: that is the part I would want checked first.

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

3 likes in reply to #33 19mo
BM
buffer_marginTL3Regular26 Dec 2024#135

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

1 like 19mo
AH
a.hartmannTL2 Moderator26 Dec 2024#136

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes 19mo

Suggested topics

TopicParticipantsRepliesViewsActivity
Validating a method you did not develop
On the subject in the title: Validating a method you did not develop Working notes rather than a conclusion. I would like to understand what this number means before I repeat it anywhere. A VendorInvestigate…
CRSOSSSVJW+10 14 15k 3mo
Revisiting: Spike-and-recover accuracy for a peptide assay
Revisiting: Spike-and-recover accuracy for a peptide assay — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I repeat it anywhere.…
SCVKEC 2 6.8k 21h
Why one laboratory's 98.4% and another's 96.1% may not contradict — one year on
Asking directly, because I could not find a straight answer: Why one laboratory's 98.4% and another's 96.1% may not contradict — one year on Working through the identity arithmetic and I would like it…
MSFKJBMRCR+10 14 13k 5mo
Revisiting: Sample handling between despatch and injection
Posting this under the heading it deserves: Revisiting: Sample handling between despatch and injection Everything below is what sits behind that. Posting the method first, because I know what the first three…
DVEAGTEMNG+5 9 1.1k 6mo
LOD and LOQ: definitions people get wrong
LOD and LOQ: definitions people get wrong — setting out what I have, and where I think it stops being reliable. A documentation question rather than an analytical one. I have a certificate in front of me that…
EISFCN 2 43k 8mo

Related topics — sharing the tags precision & repeatability, LOD & LOQ, linearity

TopicParticipantsRepliesViewsActivity
Specificity, linearity, accuracy, precision, range, robustness — with real criteria
Specificity, linearity, accuracy, precision, range, robustness — with real criteria Writing it up because I had to work it out twice and would rather nobody else did. I would like to understand what this…
FNISMLAWM+44 48 12k 1d
A comparison of home results against a laboratory result on the same vial — does this still hold?
A comparison of home results against a laboratory result on the same vial — does this still hold? I have a specific reason for asking rather than idle curiosity, and the context is below. Working through the…
BKZSCPFYLP+2 6 15k 11h
Validating a method you did not develop
On the subject in the title: Validating a method you did not develop Working notes rather than a conclusion. I would like to understand what this number means before I repeat it anywhere. A VendorInvestigate…
CRSOSSSVJW+10 14 15k 3mo
When to suspect the method rather than the sample
When to suspect the method rather than the sample — that is the question, and I have not found it answered plainly anywhere I have looked. Posting the method first, because I know what the first three replies…
MMCCYMRMBJ+109 115 63k 15mo
Relative retention time as a better inter-laboratory comparator — does this still hold?
Relative retention time as a better inter-laboratory comparator — does this still hold? I have a specific reason for asking rather than idle curiosity, and the context is below. A documentation question…
EARFWNFWOO+11 15 21k 21mo