The Peptide CommonsEst. May 2024
Independent. We sell nothing and are affiliated with no manufacturer or pharmacy. Every moderation action is logged in public
Analytics · Impurities & related substances

Follow-up: Attributing a new peak: degradation product or excipient?

Solved
Solved by st.diallo in post #6
post #5 is right about the mechanism and I think understates the practical bit. Residual solvents: traces of solvents used in purification. These are usually tested by gas chromatography, not by HPLC. A specification for residual solvents should be stated separately from the purity.

Jump to the accepted answer →

K
KStephanopoulosTL3Regular13 Feb 2025#1

Attributing a new peak: degradation product or excipient? I have a specific reason for asking rather than idle curiosity, and the context is below.

Posting the method first, because I know what the first three replies will otherwise be.

  • Column: C18, 4.6 x 250 mm, 5 um
  • Mobile phase: 0.1% TFA in water / 0.1% TFA in acetonitrile
  • Gradient: 11% to 54% organic over 20 minutes
  • Detection: 220 nm
  • Injection: 6 uL
  • Sample: tirzepatide, reconstituted to 2.0 mg/mL, injected within an hour

The main peak integrates at 97% of total area. There is a small feature on the trailing edge that I cannot decide is a shoulder or a baseline artefact, and that is what I am actually asking about.

0 likes 17mo
IA
i.amankwahTL2 Moderator14 Feb 2025#2

Oxidation at methionine and tryptophan: adds 16 per oxygen. Usually elutes earlier. Oxidation is common in storage, especially if the solution is exposed to light or if antioxidants are not present.

18 likes 17mo
B
BBramleyTL3Regular14 Feb 2025#3

Coming back to the opening post, because the follow-up matters more than the original answer.

Aggregates: multiples of the monomer mass. May not elute at all under a standard reversed-phase method. A species that does not come off the column does not appear in the area percentage.

4 likes 17mo
CS
c.serranoTL2 Moderator14 Feb 2025#4

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

0 likes 17mo
GV
g.valckenaereTL3Regular15 Feb 2025#5

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

0 likes 17mo
SD
st.dialloTL2 Moderator Solution15 Feb 2025#6

post #5 is right about the mechanism and I think understates the practical bit.

Residual solvents: traces of solvents used in purification. These are usually tested by gas chromatography, not by HPLC. A specification for residual solvents should be stated separately from the purity.

25 likes 17mo
JV
j.vandermolenTL3Regular15 Feb 2025 · edited#7
KStephanopoulos, post #1: Attributing a new peak: degradation product or excipient? I have a specific reason for asking rather than idle curiosity, and the context is below. Posting the method first, because I know what the first three replies will otherwise be. Column: C18, 4.6 x 250 mm, 5 um Mobile phase: 0.1% TFA in water / 0.1% TFA in acetonitrile Gradient:… Go to post

Disulfide formation: if a peptide contains cysteine, it can form disulfide bonds with itself or with other molecules. Under oxidising conditions multiple species appear. Reducing conditions (like DTT) convert them back.

7 likes in reply to #1 17mo
AL
a.lindqvistTL2 Moderator16 Feb 2025#8

Related substances: compounds chemically related to the target peptide but not the target peptide itself. The standard method separates them and reports them as area percent. How related they can be before they exceed specification is a regulatory question.

1 like 17mo
K
KAnderssonTL3Regular16 Feb 2025#9
a.lindqvist, post #8: Related substances: compounds chemically related to the target peptide but not the target peptide itself. The standard method separates them and reports them as area percent. How related they can be before they exceed specification is a regulatory question. Go to post

On post #5 — agreed on the reasoning, with one qualification.

Acetate content: counter-ion content. Trifluoroacetate or acetate from the salt form of the peptide. Affects mass calculations and should be stated on a complete certificate.

0 likes in reply to #8 17mo
MI
m.ilungaTL2 Moderator16 Feb 2025#10
j.vandermolen, post #7: Disulfide formation: if a peptide contains cysteine, it can form disulfide bonds with itself or with other molecules. Under oxidising conditions multiple species appear. Reducing conditions (like DTT) convert them back. Go to post

post #9 answers the question as asked. The question underneath it is different.

Dimer and higher-order multimers: two or more peptide molecules bonded together. They appear at double the mass and higher. They may or may not separate from the monomer on HPLC depending on the method.

33 likes in reply to #7 17mo
JW
journalclub_wrenTL3Regular17 Feb 2025 · edited#11
g.valckenaere, post #5: Two things before anyone answers the substance. First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound. Go to post

post #10 is right about the mechanism and I think understates the practical bit.

Truncation products: fragments from incomplete synthesis or from degradation. They elute quite differently from the intact peptide because they are much smaller and have different hydrophobicity. They are usually well separated.

0 likes in reply to #5 17mo
NC
n.cabreraTL2 Moderator17 Feb 2025#12
a.lindqvist, post #8: Related substances: compounds chemically related to the target peptide but not the target peptide itself. The standard method separates them and reports them as area percent. How related they can be before they exceed specification is a regulatory question. Go to post

Worth separating two things that post #8 runs together.

Off-target structures: if the sequence synthesis goes wrong, a completely different amino acid can be incorporated. The resulting off-target peptide is a structural isomer with the same mass but a different sequence. No chromatographic purity method detects this without a reference standard.

0 likes in reply to #8 17mo
SS
steady_stateTL3Regular17 Feb 2025#13

Deletion sequences (incomplete coupling during synthesis): lower in mass by one residue. Chromatographically they usually elute earlier or later depending on the residue's hydrophobicity. They are the most common impurity in solid-phase synthesis.

8 likes 17mo
SV
s.vukovicTL2 Moderator17 Feb 2025#14

Incomplete deprotection: mass higher by the protecting group mass. Usually markedly later eluting. A synthesis artifact from incomplete removal of protecting groups.

19 likes 17mo
NA
n.abernathyTL3Analytical chemist18 Feb 2025#15
st.diallo, post #6: post #5 is right about the mechanism and I think understates the practical bit. Residual solvents: traces of solvents used in purification. These are usually tested by gas chromatography, not by HPLC. A specification for residual solvents should be stated separately from the purity. Go to post

Off-target structures: if the sequence synthesis goes wrong, a completely different amino acid can be incorporated. The resulting off-target peptide is a structural isomer with the same mass but a different sequence. No chromatographic purity method detects this without a reference standard.

0 likes in reply to #6 17mo
HA
h.agyemanTL2 Moderator18 Feb 2025#16

On post #12 — agreed on the reasoning, with one qualification.

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

2 likes 17mo
DH
dietitian_hollisTL3Dietitian18 Feb 2025#17

Dimer and higher-order multimers: two or more peptide molecules bonded together. They appear at double the mass and higher. They may or may not separate from the monomer on HPLC depending on the method.

12 likes 17mo
BK
b.kowalskiTL2 Moderator18 Feb 2025#18

Truncation products: fragments from incomplete synthesis or from degradation. They elute quite differently from the intact peptide because they are much smaller and have different hydrophobicity. They are usually well separated.

26 likes 17mo
CT
cannula_traceTL3Regular19 Feb 2025#19
st.diallo, post #6: post #5 is right about the mechanism and I think understates the practical bit. Residual solvents: traces of solvents used in purification. These are usually tested by gas chromatography, not by HPLC. A specification for residual solvents should be stated separately from the purity. Go to post

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

20 likes in reply to #6 17mo
VR
v.rautioTL2 Moderator19 Feb 2025 · edited#20

Deletion sequences (incomplete coupling during synthesis): lower in mass by one residue. Chromatographically they usually elute earlier or later depending on the residue's hydrophobicity. They are the most common impurity in solid-phase synthesis.

0 likes 17mo
Moved from Mass spectrometry by dr_okonkwo. Category placement is not obvious from outside and getting it wrong is expected. This topic will get better answers here. The move is recorded in the public log citing R7.

Suggested topics

TopicParticipantsRepliesViewsActivity
Attributing a new peak: degradation product or excipient?
Asking directly, because I could not find a straight answer: Attributing a new peak: degradation product or excipient? Posting the method first, because I know what the first three replies will otherwise be.…
SHOWM 2 15k 18mo
Deletion sequences: expected mass shifts by residue
Deletion sequences: expected mass shifts by residue — setting out what I have, and where I think it stops being reliable. Working through the identity arithmetic and I would like it checked. semaglutide has a…
SCBBCN 2 59 3mo
Racemisation and why it is invisible to a mass measurement — what changed since
Racemisation and why it is invisible to a mass measurement — what changed since Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I…
TVBVCCBIN+11 15 42k 16d
Trifluoroacetate content and its consequences
Trifluoroacetate content and its consequences — setting out what I have, and where I think it stops being reliable. A documentation question rather than an analytical one. I have a certificate in front of me…
AIRDRVHFAV+123 134 31k 5mo
Endotoxin and why it is a separate question from purity — what changed since
On the subject in the title: Endotoxin and why it is a separate question from purity — what changed since Working notes rather than a conclusion. A documentation question rather than an analytical one. I have…
JSPRMVMDEM+78 91 1k 10h

Related topics — sharing the tags water content, HPLC, forced degradation

TopicParticipantsRepliesViewsActivity
Racemisation and why it is invisible to a mass measurement
Posting this under the heading it deserves: Racemisation and why it is invisible to a mass measurement Everything below is what sits behind that. Working through the identity arithmetic and I would like it…
NTKCTDOPA+120 127 41k 1mo
Second pass at: What a 2 ppm match establishes, and what it does not
Second pass at: What a 2 ppm match establishes, and what it does not — setting out what I have, and where I think it stops being reliable. Working through the identity arithmetic and I would like it checked.…
BSNKSTMGMM+44 49 8.1k 1d
Detection wavelength and why 214 nm and 280 nm disagree
On the subject in the title: Detection wavelength and why 214 nm and 280 nm disagree Working notes rather than a conclusion. Posting the method first, because I know what the first three replies will…
BWZVGFFEDI+78 84 1.7k 8mo
Carryover and the ghost peak from last week's standard — what changed since
Carryover and the ghost peak from last week's standard — what changed since Writing it up because I had to work it out twice and would rather nobody else did. Posting the method first, because I know what the…
FDLTZYFWAA+98 114 37k 22d
Where impurities in solid-phase peptide synthesis come from
Where impurities in solid-phase peptide synthesis come from Writing it up because I had to work it out twice and would rather nobody else did. A documentation question rather than an analytical one. I have a…
LCDSPOCLNK+54 59 14k 3mo