MS/MS sequence confirmation: what it costs and when it is worth it — what changed since

Closed
AC
a.cabreraTL2 Moderator2 Feb 2026#1

Posting this under the heading it deserves: MS/MS sequence confirmation: what it costs and when it is worth it — what changed since Everything below is what sits behind that.

Working through the identity arithmetic and I would like it checked.

tirzepatide has a monoisotopic mass close to 4813.5 Da. On an electrospray instrument I would expect to see the multiply charged series rather than the intact singly charged ion, so for the doubly charged species I calculate (4813.5 + 2 x 1.00728) / 2, and for the triply charged the analogous expression.

The observed values in the report sit within a few ppm of those. My question is what that actually establishes, because I have seen people treat a mass match as a purity result and I do not think it is one.

16 likes 6mo
CC
crossref_checkTL3Wiki editor10 Feb 2026 · edited#2

On the opening post — agreed on the reasoning, with one qualification.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

20 likes 6mo
ND
n.duarteTL2 Moderator16 Feb 2026#3
a.cabrera, post #1: Posting this under the heading it deserves: MS/MS sequence confirmation: what it costs and when it is worth it — what changed since Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked. tirzepatide has a monoisotopic mass close to 4813.5 Da. On an electrospray instrument I would… Go to post

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

0 likes in reply to #1 5mo
V
VPoulsenTL3Regular21 Feb 2026#4

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

2 likes 5mo
VB
v.bergstromTL2 Moderator26 Feb 2026#5

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

5 likes 5mo
RJ
r.jhannsdttirTL3Regular2 Mar 2026#6

Worth separating two things that post #2 runs together.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

14 likes 5mo
AV
a.vermeulenTL2 Moderator6 Mar 2026#7
crossref_check, post #2: On the opening post — agreed on the reasoning, with one qualification. Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

This follows post #4 rather than contradicting it.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes in reply to #2 5mo
TK
t.kulkarniTL3Regular10 Mar 2026#8

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes 5mo
GO
g.oyelaranTL2 Moderator14 Mar 2026#9

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

19 likes 4mo
HA
h.almeidaTL2Member18 Mar 2026#10
n.duarte, post #3: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

0 likes in reply to #3 4mo
B
BBramleyTL3Regular22 Mar 2026#11
VPoulsen, post #4: What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

6 likes in reply to #4 4mo
CS
c.serranoTL2 Moderator26 Mar 2026#12

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

1 like 4mo
TN
t.nardoneTL3Regular29 Mar 2026#13

On post #9 — agreed on the reasoning, with one qualification.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 4mo
IA
i.amankwahTL2 Moderator2 Apr 2026#14
g.oyelaran, post #9: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

22 likes in reply to #9 4mo
LC
l.chevalierTL3Regular5 Apr 2026#15
t.nardone, post #13: On post #9 — agreed on the reasoning, with one qualification. Resolution: "high resolution" commonly means Go to post

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

3 likes in reply to #13 4mo
EM
e.mensaTL2 Moderator9 Apr 2026#16

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

0 likes 4mo
VD
vial_deskTL3Regular12 Apr 2026#17

Worth separating two things that post #13 runs together.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

31 likes 4mo
AE
a.eriksenTL2 Moderator16 Apr 2026 · edited#18

post #17 is right about the mechanism and I think understates the practical bit.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

16 likes 3mo
EC
excursion_checkTL3Regular19 Apr 2026#19
l.chevalier, post #15: Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity. Go to post

Coming back to post #17, because the follow-up matters more than the original answer.

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

1 like in reply to #15 3mo
SV
s.vanheckeTL2 Moderator22 Apr 2026#20
BBramley, post #11: Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency. Go to post

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes in reply to #11 3mo
MP
mira.patelTL4 Admin25 Apr 2026#21

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

3 likes 3mo
EV
e.vargaTL2 Moderator28 Apr 2026#22
h.almeida, post #10: Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity. Go to post

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

11 likes in reply to #10 3mo
AA
a.adebayoTL2 Moderator2 May 2026 · edited#23
vial_desk, post #17: Worth separating two things that post #13 runs together. Resolution: "high resolution" commonly means Go to post

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

32 likes in reply to #17 3mo
AN
a.novakTL25 May 2026#24
DT
d.tammTL2 Moderator8 May 2026#25

This follows post #22 rather than contradicting it.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

1 like 3mo
BF
b.fonsecaTL2 Moderator11 May 2026#26
g.oyelaran, post #9: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

7 likes in reply to #9 3mo
CS
c.silvaTL2 Moderator14 May 2026#27

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

24 likes 2mo
AZ
a.zamoraTL2 Moderator17 May 2026#28

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

0 likes 2mo
EF
e.ferreiraTL3Regular20 May 2026#29

Picking up post #26: that is the part I would want checked first.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

10 likes 2mo
AJ
a.jansenTL2 Moderator23 May 2026#30

Coming back to post #28, because the follow-up matters more than the original answer.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.