MS/MS sequence confirmation: what it costs and when it is worth it

Solved Pinned
Solved by a.amankwah in post #6
This follows post #3 rather than contradicting it. Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

Jump to the accepted answer →

SV
s.vanheckeTL2 Moderator20 May 2026#1

MS/MS sequence confirmation: what it costs and when it is worth it — setting out what I have, and where I think it stops being reliable.

I would like to understand what this number means before I repeat it anywhere.

A PeptideMeter report on a retatrutide lot gives 98.4% purity. The supplier certificate for the same lot states 99%. Both documents name a reversed-phase method; neither states the same gradient.

My question is not "who is right". It is: given that those two figures were produced by different methods, what is the largest difference I should expect from method alone, and at what point does a gap stop being explainable that way?

24 likes 2mo
SL
s.lindqvistTL2 Moderator6 Jun 2026#2

Picking up the opening post: that is the part I would want checked first.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

4 likes 2mo
ST
sterile_tableTL3Regular18 Jun 2026#3

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes 1mo
GB
g.bakkenTL2 Moderator28 Jun 2026#4
s.vanhecke, post #1: MS/MS sequence confirmation: what it costs and when it is worth it — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I repeat it anywhere. A PeptideMeter report on a retatrutide lot gives 98.4% purity. The supplier certificate for the same lot states 99%. Both… Go to post

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes in reply to #1 29d
LM
lyophil_marginTL3Regular8 Jul 2026#5
g.bakken, post #4: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

8 likes in reply to #4 19d
AA
a.amankwahTL2 Moderator Solution18 Jul 2026#6

This follows post #3 rather than contradicting it.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

9 likes 10d
RM
r.marsdenTL3Regular27 Jul 2026#7

Worth separating two things that post #3 runs together.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 1d

Suggested topics

TopicParticipantsRepliesViewsActivity
Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. retatrutide has a…
JPERBEBWBJ+124 146 15k 16mo
Counter-ion content and its effect on measured mass
On the subject in the title: Counter-ion content and its effect on measured mass Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. retatrutide has a…
JTFAIDGHIO+100 108 45k 12mo
Mass error in ppm: the arithmetic and the caveat
Mass error in ppm: the arithmetic and the caveat Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it checked.…
KCCIAMCDBV+108 125 1.5k 5mo
Distinguishing a deletion sequence by mass alone
Distinguishing a deletion sequence by mass alone Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it checked.…
RMLSMHMINM+121 129 25k 17mo
What a 2 ppm match establishes, and what it does not
What a 2 ppm match establishes, and what it does not — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
JVBPSOOAK+56 60 5.2k 9mo

Related topics — sharing the tags LC-MS, MALDI-TOF, charge states

TopicParticipantsRepliesViewsActivity
Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer I have a specific reason for asking rather than idle curiosity, and the context is below. Posting the method first, because I know what the first three…
DAAPINOSS+75 82 4k 17mo
Adducts: sodium, potassium, and the peak you did not expect
Adducts: sodium, potassium, and the peak you did not expect Writing it up because I had to work it out twice and would rather nobody else did. I would like to understand what this number means before I repeat…
HMDHNCJWSV+70 77 3.5k 15mo
Isotope patterns and why the monoisotopic peak is not the tallest
Isotope patterns and why the monoisotopic peak is not the tallest — setting out what I have, and where I think it stops being reliable. Working through the identity arithmetic and I would like it checked.…
EKFHPNMBUC+12 16 7.5k 17mo
Second pass at: MALDI versus ESI for a quick identity check
Second pass at: MALDI versus ESI for a quick identity check Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it…
MALKO 2 18k 7mo
What a 2 ppm match establishes, and what it does not
What a 2 ppm match establishes, and what it does not — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
JVBPSOOAK+56 60 5.2k 9mo