System suitability testing: injections run before and during the sample run to establish whether the instrument, column and method were performing when the sample was analysed. If suitability did not pass, the sample results from that run are uninterpretable.
Coming back to: Resolution and tailing factor: real acceptance criteria
This follows post #35 rather than contradicting it.
Reversed-phase separates on hydrophobicity. A peptide is retained on a non-polar stationary phase and eluted by increasing organic solvent. For peptides the mobile phase almost always contains an ion-pairing acid, typically 0.1% TFA, which suppresses secondary interactions and sharpens peaks.
Worth separating two things that post #60 runs together.
Before anything else: what was the gradient, and at what wavelength? Area percent at different wavelengths is not the same number even on the same sample because different species absorb differently at different wavelengths. With the method stated, I can tell you something useful. Without it, all I can say is that there is one large peak.
Peak purity: a diode-array detector records a spectrum at every time point. If a peak contains two co-eluting species with different spectra, the spectrum changes across the peak. A passing peak-purity result says the spectrum is constant; it is weak evidence of homogeneity if the impurities have similar spectra.
I disagree with the reply above, and I think the disagreement is substantive rather than terminological.
The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.
For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.
This follows post #93 rather than contradicting it.
Before anything else: what was the gradient, and at what wavelength? Area percent at different wavelengths is not the same number even on the same sample because different species absorb differently at different wavelengths. With the method stated, I can tell you something useful. Without it, all I can say is that there is one large peak.
On post #105 — agreed on the reasoning, with one qualification.
Method validation is the demonstration that a method can separate the compound from its degradation products and impurities reliably. A method that cannot resolve an impurity from the parent peak will not detect that impurity.
Read the full topic (113 posts)
Suggested topics
| Topic | Participants | Replies | Views | Activity |
|---|---|---|---|---|
|
Injection volume, overload, and peak distortion — one year on
Injection volume, overload, and peak distortion — one year on Writing it up because I had to work it out twice and would rather nobody else did. A documentation question rather than an analytical one. I have…
|
+4 | 8 | 1.2k | 4mo |
|
Revisiting: Theoretical plates: useful number or ritual?
Asking directly, because I could not find a straight answer: Revisiting: Theoretical plates: useful number or ritual? I would like to understand what this number means before I repeat it anywhere. A…
|
+16 | 21 | 1.2k | 1d |
|
[2026 update] Detection wavelength and why 214 nm and 280 nm disagree
Detection wavelength and why 214 nm and 280 nm disagree — setting out what I have, and where I think it stops being reliable. Posting the method first, because I know what the first three replies will…
|
+69 | 76 | 920 | 6h |
|
Gradient steepness and apparent purity: a worked comparison
On the subject in the title: Gradient steepness and apparent purity: a worked comparison Working notes rather than a conclusion. A documentation question rather than an analytical one. I have a certificate in…
|
2 | 51k | 22mo | |
|
What a reversed-phase purity number actually is — the long version
What a reversed-phase purity number actually is — the long version I have a specific reason for asking rather than idle curiosity, and the context is below. I would like to understand what this number means…
|
2 | 15k | 8mo |
Related topics — sharing the tags peak integration, data table, UV & DAD detection
| Topic | Participants | Replies | Views | Activity |
|---|---|---|---|---|
|
Carryover and the ghost peak from last week's standard
On the subject in the title: Carryover and the ghost peak from last week's standard Working notes rather than a conclusion. Posting the method first, because I know what the first three replies will otherwise…
|
+80 | 92 | 16k | 16mo |
|
Blinding yourself: practical methods and their limits
On the subject in the title: Blinding yourself: practical methods and their limits Working notes rather than a conclusion. Comparing PIONEER 6 ( N Engl J Med , 2019) with STEP 4 ( JAMA , 2021) and finding the…
|
+40 | 44 | 33k | 5d |
|
Revisiting: Sample size in a voluntary survey: the selection problem
Revisiting: Sample size in a voluntary survey: the selection problem — setting out what I have, and where I think it stops being reliable. Posting the method first, because I know what the first three replies…
|
2 | 56k | 9mo | |
|
Follow-up: Interpreting a composition change that is within the error bars
Interpreting a composition change that is within the error bars Writing it up because I had to work it out twice and would rather nobody else did. Longitudinal report with the method stated, because a number…
|
2 | 95 | 24d | |
|
Deamidation and the close-eluting pair it produces — the long version
Posting this under the heading it deserves: Deamidation and the close-eluting pair it produces — the long version Everything below is what sits behind that. I would like to understand what this number means…
|
3 | 347 | 24d |