Repeatability versus intermediate precision

Solved
Solved by a.ibarra in post #3
On the opening post — agreed on the reasoning, with one qualification. Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

Jump to the accepted answer →

LS
l.sarkissianTL2Member15 Jun 2025#1

Repeatability versus intermediate precision — setting out what I have, and where I think it stops being reliable.

Posting the method first, because I know what the first three replies will otherwise be.

  • Column: C8, 4.6 x 150 mm, 3.5 um
  • Mobile phase: 0.1% TFA in water / 0.1% TFA in acetonitrile
  • Gradient: 5% to 62% organic over 26 minutes
  • Detection: 220 nm
  • Injection: 12 uL
  • Sample: tirzepatide, reconstituted to 0.5 mg/mL, injected within an hour

The main peak integrates at 96.4% of total area. There is a small feature on the trailing edge that I cannot decide is a shoulder or a baseline artefact, and that is what I am actually asking about.

15 likes 13mo
K
KLindqvistTL4 Moderator16 Jun 2025#2
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

2 likes 13mo
AI
a.ibarraTL2 Moderator Solution16 Jun 2025 · edited#3
l.sarkissian, post #1: Repeatability versus intermediate precision — setting out what I have, and where I think it stops being reliable. Posting the method first, because I know what the first three replies will otherwise be. Column: C8, 4.6 x 150 mm, 3.5 um Mobile phase: 0.1% TFA in water / 0.1% TFA in acetonitrile Gradient: 5% to 62% organic over 26 minutes… Go to post

On the opening post — agreed on the reasoning, with one qualification.

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

7 likes in reply to #1 13mo
LW
l.wikstromTL2 Moderator16 Jun 2025#4

post #3 answers the question as asked. The question underneath it is different.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

26 likes 13mo
NO
n.oseiTL217 Jun 2025#5
BD
baseline_driftTL2Analytical chemist17 Jun 2025#6

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

0 likes 13mo
NK
n.krastevTL2 Moderator17 Jun 2025#7
l.wikstrom, post #4: post #3 answers the question as asked. The question underneath it is different. Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

0 likes in reply to #4 13mo
DO
d.oyelaranTL3Pharmacist17 Jun 2025#8
a.ibarra, post #3: On the opening post — agreed on the reasoning, with one qualification. Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements. Go to post

post #7 is right about the mechanism and I think understates the practical bit.

Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99.

19 likes in reply to #3 13mo
SS
s.silvaTL2 Moderator17 Jun 2025#9

Coming back to post #7, because the follow-up matters more than the original answer.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

0 likes 13mo
JH
j.hartmannTL2 Moderator18 Jun 2025 · edited#10
baseline_drift, post #6: Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount. Go to post

Picking up post #7: that is the part I would want checked first.

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

25 likes in reply to #6 13mo
FF
f.fontaineTL2 Moderator18 Jun 2025#11

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

1 like 13mo
RM
r.marsdenTL3Regular18 Jun 2025#12

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

6 likes 13mo
SR
s.roosTL2 Moderator18 Jun 2025#13
r.marsden, post #12: System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5). Go to post

post #12 answers the question as asked. The question underneath it is different.

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

16 likes in reply to #12 13mo
GV
g.valckenaereTL3Regular18 Jun 2025#14

On post #10 — agreed on the reasoning, with one qualification.

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

31 likes 13mo
MR
m.ramosTL2 Moderator19 Jun 2025#15

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

0 likes 13mo
DN
desiccant_notesTL2Member19 Jun 2025 · edited#16

I read post #14 twice before replying, because I had assumed the opposite.

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

3 likes 13mo
JI
j.ivaturiTL2 Moderator19 Jun 2025#17
l.sarkissian, post #1: Repeatability versus intermediate precision — setting out what I have, and where I think it stops being reliable. Posting the method first, because I know what the first three replies will otherwise be. Column: C8, 4.6 x 150 mm, 3.5 um Mobile phase: 0.1% TFA in water / 0.1% TFA in acetonitrile Gradient: 5% to 62% organic over 26 minutes… Go to post

post #16 is right about the mechanism and I think understates the practical bit.

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

11 likes in reply to #1 13mo
BO
b.oylerTL1Member19 Jun 2025#18

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

24 likes 13mo
EM
e.mbekiTL2 Moderator19 Jun 2025#19

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

0 likes 13mo
RH
revision_historyTL3Wiki editor19 Jun 2025#20

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

1 like 13mo
IA
id.almeidaTL2 Moderator20 Jun 2025#21

Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99.

0 likes 13mo
BD
baseline_driftTL2Analytical chemist20 Jun 2025#22
revision_history, post #20: Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements. Go to post

Picking up post #19: that is the part I would want checked first.

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

0 likes in reply to #20 13mo
NK
n.krastevTL2 Moderator20 Jun 2025#23

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

14 likes 13mo
DO
d.oyelaranTL3Pharmacist20 Jun 2025#24

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

5 likes 13mo
SK
s.kuuselaTL2 Moderator20 Jun 2025#25

I read post #23 twice before replying, because I had assumed the opposite.

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

2 likes 13mo
B
batchlogTL3Regular20 Jun 2025#26
l.sarkissian, post #1: Repeatability versus intermediate precision — setting out what I have, and where I think it stops being reliable. Posting the method first, because I know what the first three replies will otherwise be. Column: C8, 4.6 x 150 mm, 3.5 um Mobile phase: 0.1% TFA in water / 0.1% TFA in acetonitrile Gradient: 5% to 62% organic over 26 minutes… Go to post

This follows post #23 rather than contradicting it.

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

0 likes in reply to #1 13mo
JI
j.iyerTL2 Moderator20 Jun 2025#27
baseline_drift, post #22: Picking up post #19: that is the part I would want checked first. Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal). Go to post

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

20 likes in reply to #22 13mo
BV
bias_varianceTL4Biostatistician21 Jun 2025 · edited#28

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

9 likes 13mo
VN
v.nascimentoTL2 Moderator21 Jun 2025#29

Coming back to post #27, because the follow-up matters more than the original answer.

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

5 likes 13mo
NR
n.rahimiTL2 Moderator21 Jun 2025#30

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

0 likes 13mo