Spike-and-recover accuracy for a peptide assay posts 91–120

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1.

MA
m.almeidaTL2 Moderator10 Apr 2026#91

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

1 like 4mo
AK
a.kowalczykTL2Regular10 Apr 2026#92

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

0 likes 4mo
CH
c.haddadTL2 Moderator10 Apr 2026#93
m.haddad, post #85: Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong. Go to post

On post #89 — agreed on the reasoning, with one qualification.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

25 likes in reply to #85 4mo
PN
plateau_notesTL210 Apr 2026#94
AK
a.krastevTL2 Moderator10 Apr 2026#95

I read post #93 twice before replying, because I had assumed the opposite.

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

0 likes 4mo
DB
d.bramleyTL3Regular10 Apr 2026#96

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

0 likes 4mo
AN
a.nascimentoTL2 Moderator10 Apr 2026#97

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

18 likes 4mo
KB
k.brandl_deTL3Translator · DE10 Apr 2026#98
d.ndiaye, post #58: post #57 answers the question as asked. The question underneath it is different. System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution… Go to post

post #97 is right about the mechanism and I think understates the practical bit.

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

7 likes in reply to #58 4mo
RP
r.petrovTL2 Moderator10 Apr 2026 · edited#99

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

7 likes 4mo
PE
ppm_errorTL3Analytical chemist10 Apr 2026#100

Picking up post #97: that is the part I would want checked first.

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

1 like 4mo
LF
l.ferreiraTL2 Moderator10 Apr 2026#101

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

0 likes 4mo
HK
h.koodziejTL2Member10 Apr 2026#102

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

30 likes 4mo
AA
an.adeyemiTL2 Moderator10 Apr 2026#103

Coming back to post #101, because the follow-up matters more than the original answer.

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

15 likes 4mo
ES
e.silvaTL2 Moderator10 Apr 2026#104
a.iyer, post #36: System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5). Go to post

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

5 likes in reply to #36 4mo
CR
c.rasmussenTL2 Moderator10 Apr 2026 · edited#105

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

0 likes 4mo
EL
e.lokkenTL2 Moderator10 Apr 2026#106

post #105 is right about the mechanism and I think understates the practical bit.

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

22 likes 4mo
JC
j.castellanosTL2 Moderator10 Apr 2026#107

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

10 likes 4mo
JS
j.sorensenTL2 Moderator10 Apr 2026#108
c.inglethorpe, post #26: I read post #24 twice before replying, because I had assumed the opposite. Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a… Go to post

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

3 likes in reply to #26 4mo
SL
s.lundgrenTL2 Moderator10 Apr 2026#109

On post #105 — agreed on the reasoning, with one qualification.

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

2 likes 4mo
AS
a.stephanopoulosTL3Regular10 Apr 2026#110

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

0 likes 4mo
JL
j.lokkenTL2 Moderator10 Apr 2026 · edited#111

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

1 like 4mo
SS
s.stavrianosTL2Member10 Apr 2026#112
apostille_trace, post #78: Picking up post #75: that is the part I would want checked first. Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples. Go to post

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

6 likes in reply to #78 4mo
LK
l.krastevTL2 Moderator10 Apr 2026#113

post #112 answers the question as asked. The question underneath it is different.

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

16 likes 4mo
D
DKwiatkowskiTL3Regular10 Apr 2026#114

On post #110 — agreed on the reasoning, with one qualification.

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

31 likes 4mo
AV
a.vestergaardTL210 Apr 2026#115
GD
glossary_deskTL3Regular11 Apr 2026#116
bias_variance, post #31: On post #27 — agreed on the reasoning, with one qualification. Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements. Go to post

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

3 likes in reply to #31 4mo
VS
v.stanescuTL2 Moderator11 Apr 2026#117

post #116 is right about the mechanism and I think understates the practical bit.

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

11 likes 4mo
NM
n.marsdenTL1Member11 Apr 2026#118

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

23 likes 4mo
JS
j.solbergTL2 Moderator11 Apr 2026#119

Picking up post #116: that is the part I would want checked first.

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

6 likes 4mo
L
LundqvistTL2Member11 Apr 2026#120

Coming back to post #118, because the follow-up matters more than the original answer.

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.