Spike-and-recover accuracy for a peptide assay posts 31–60

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1.

BV
bias_varianceTL4Biostatistician7 Apr 2026 · edited#31

On post #27 — agreed on the reasoning, with one qualification.

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

15 likes 4mo
JM
j.moreauTL2 Moderator8 Apr 2026#32

post #31 answers the question as asked. The question underneath it is different.

Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99.

5 likes 4mo
FD
f.demirTL2Regular8 Apr 2026#33

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

0 likes 4mo
MS
m.steinerTL2 Moderator8 Apr 2026#34
h.amankwah, post #21: Picking up post #18: that is the part I would want checked first. Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance. Go to post

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

29 likes in reply to #21 4mo
RM
r.mcalisterTL3Regular8 Apr 2026#35
h.nwosu, post #18: Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Worth separating two things that post #31 runs together.

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

21 likes in reply to #18 4mo
AI
a.iyerTL2 Moderator8 Apr 2026#36

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

9 likes 4mo
DT
dexa_twice_yearlyTL3Regular8 Apr 2026#37

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

0 likes 4mo
RN
r.nakamuraTL2 Moderator8 Apr 2026#38
CSagredo, post #28: Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing. The correction was fair and I had been repeating something I had not checked carefully enough. Go to post

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

0 likes in reply to #28 4mo
SC
so.cardosoTL2 Moderator8 Apr 2026#39

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

28 likes 4mo
SD
s.dziedzicTL2 Moderator8 Apr 2026#40

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

14 likes 4mo
AP
abstract_peakTL1Member8 Apr 2026#41

post #40 answers the question as asked. The question underneath it is different.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

25 likes 4mo
NA
n.achebeTL2 Moderator8 Apr 2026#42
peak_purity, post #7: I disagree with the reply above, and I think the disagreement is substantive rather than terminological. The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient. Go to post

On post #38 — agreed on the reasoning, with one qualification.

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

0 likes in reply to #7 4mo
TN
t.nardoneTL3Regular8 Apr 2026#43

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

1 like 4mo
TI
t.ibarraTL2 Moderator8 Apr 2026#44

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

7 likes 4mo
TT
titrate_traceTL1Member8 Apr 2026#45

post #44 is right about the mechanism and I think understates the practical bit.

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

0 likes 4mo
RB
r.bakkenTL2 Moderator8 Apr 2026 · edited#46
j.cabrera, post #12: Picking up post #9: that is the part I would want checked first. Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that… Go to post

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

0 likes in reply to #12 4mo
NR
n.rowntreeTL3Regular8 Apr 2026#47

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

4 likes 4mo
BC
b.correiaTL2 Moderator8 Apr 2026#48

I read post #46 twice before replying, because I had assumed the opposite.

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

12 likes 4mo
DS
d.szymanskiTL3Wiki editor8 Apr 2026#49
peak_purity, post #7: I disagree with the reply above, and I think the disagreement is substantive rather than terminological. The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient. Go to post

Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99.

0 likes in reply to #7 4mo
EN
e.ndiayeTL2 Moderator8 Apr 2026#50
h.nwosu, post #18: Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

1 like in reply to #18 4mo
MS
m.stephanopoulosTL3Regular8 Apr 2026#51

I read post #49 twice before replying, because I had assumed the opposite.

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

17 likes 4mo
BN
b.nwosuTL2 Moderator8 Apr 2026#52
j.cabrera, post #12: Picking up post #9: that is the part I would want checked first. Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that… Go to post

This follows post #49 rather than contradicting it.

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

7 likes in reply to #12 4mo
BP
baseline_peakTL2Member8 Apr 2026#53

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

1 like 4mo
BJ
b.jansenTL2 Moderator8 Apr 2026#54

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

0 likes 4mo
Z
ZieglerTL3Regular8 Apr 2026 · edited#55

Coming back to post #53, because the follow-up matters more than the original answer.

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

23 likes 4mo
MD
m.dumitruTL2 Moderator8 Apr 2026#56
h.nwosu, post #18: Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

11 likes in reply to #18 4mo
W
WickramasingheTL2Member9 Apr 2026#57

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

3 likes 4mo
DN
d.ndiayeTL2 Moderator9 Apr 2026#58

post #57 answers the question as asked. The question underneath it is different.

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

0 likes 4mo
LA
l.aaltonenTL3Regular9 Apr 2026#59
r.nakamura, post #38: I disagree with the reply above, and I think the disagreement is substantive rather than terminological. The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient. Go to post

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

31 likes in reply to #38 4mo
PO
p.onwukaTL2 Moderator9 Apr 2026#60
c.inglethorpe, post #26: I read post #24 twice before replying, because I had assumed the opposite. Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a… Go to post

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

16 likes in reply to #26 4mo