Adducts: sodium, potassium, and the peak you did not expect — does this still hold? posts 61–90

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1 · go to the accepted answer.

ND
n.duarteTL2 Moderator8 Mar 2026#61

post #60 answers the question as asked. The question underneath it is different.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

15 likes 5mo
V
VPoulsenTL3Regular9 Mar 2026#62

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

29 likes 5mo
IW
i.wojcikTL2 Moderator9 Mar 2026#63
p.amankwah, post #24: Resolution: "high resolution" commonly means Go to post

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

0 likes in reply to #24 5mo
BE
bench_entryTL3Regular10 Mar 2026#64

Coming back to post #62, because the follow-up matters more than the original answer.

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

3 likes 5mo
BF
b.friskTL2 Moderator11 Mar 2026#65

post #64 is right about the mechanism and I think understates the practical bit.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

21 likes 5mo
TK
t.kulkarniTL3Regular12 Mar 2026#66

Worth separating two things that post #62 runs together.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes 5mo
GO
g.oyelaranTL2 Moderator13 Mar 2026#67
l.chevalier, post #19: On post #15 — agreed on the reasoning, with one qualification. What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

0 likes in reply to #19 5mo
IL
integrator_logTL3Regular14 Mar 2026 · edited#68
e.iyer, post #34: Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

5 likes in reply to #34 4mo
CV
c.vasquezTL2 Moderator14 Mar 2026 · edited#69

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

28 likes 4mo
CL
customs_ledgerTL3Regular15 Mar 2026#70

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 4mo
JD
j.delacroixTL3Regular16 Mar 2026 · edited#71

I read post #69 twice before replying, because I had assumed the opposite.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

9 likes 4mo
TL
t.lindqvistTL2 Moderator17 Mar 2026#72
dr.villanueva, post #42: I read post #40 twice before replying, because I had assumed the opposite. Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing. The correction was fair and I had been repeating something I had not checked carefully enough. Go to post

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

2 likes in reply to #42 4mo
M
MSaarinenTL3Regular18 Mar 2026#73

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

0 likes 4mo
BB
b.brandtTL2 Moderator19 Mar 2026#74

post #73 is right about the mechanism and I think understates the practical bit.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

28 likes 4mo
PS
p.silvaTL2 Moderator19 Mar 2026#75

Coming back to post #73, because the follow-up matters more than the original answer.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

13 likes 4mo
MM
m.malinowskiTL2 Moderator20 Mar 2026#76

Picking up post #73: that is the part I would want checked first.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

5 likes 4mo
HM
h.mukherjeeTL1Member21 Mar 2026#77
m.adebayo, post #20: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes in reply to #20 4mo
IB
i.brobergTL2 Moderator22 Mar 2026#78

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

0 likes 4mo
RT
r.torrenceTL2Member23 Mar 2026#79
z.yildiz, post #2: Resolution: "high resolution" commonly means Go to post

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

19 likes in reply to #2 4mo
CK
c.kuuselaTL2 Moderator23 Mar 2026#80
i.wojcik, post #63: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

This follows post #77 rather than contradicting it.

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

8 likes in reply to #63 4mo
M
MakinenTL2Member24 Mar 2026#81

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

26 likes 4mo
ON
o.nybergTL2 Moderator25 Mar 2026#82

Coming back to post #80, because the follow-up matters more than the original answer.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes 4mo
W
WoodhouseTL2Member26 Mar 2026#83
a.silva, post #14: Picking up post #11: that is the part I would want checked first. Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

post #82 answers the question as asked. The question underneath it is different.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

4 likes in reply to #14 4mo
SR
s.radichTL2 Moderator27 Mar 2026#84

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

12 likes 4mo
K
KForsbergTL2Member27 Mar 2026#85

This follows post #82 rather than contradicting it.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

19 likes 4mo
KK
k.kimaniTL2 Moderator28 Mar 2026#86

I read post #84 twice before replying, because I had assumed the opposite.

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

0 likes 4mo
MW
m.wanjalaTL1Member29 Mar 2026 · edited#87
m.agyeman, post #52: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

2 likes in reply to #52 4mo
SH
s.hartmannTL2 Moderator30 Mar 2026#88
ms_holloway, post #44: Resolution: "high resolution" commonly means Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

8 likes in reply to #44 4mo
RF
r.friskTL2 Moderator31 Mar 2026#89

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

13 likes 4mo
HF
h.ferrariTL2 Moderator31 Mar 2026 · edited#90
d.petrescu, post #57: What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

27 likes in reply to #57 4mo