Mass error in ppm: the arithmetic and the caveat — a second dataset

This is a generated summary. It shows the 5 most-liked posts from a topic of 25, in their original order, with the accepted answer included where one exists. It is a reading aid and it will miss nuance — the full topic is the record.
ON
o.nybergTL2 Moderator29 May 2026#4

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

28 likes 2mo
CV
ca.vermeulenTL2 Moderator Solution12 Jun 2026#8

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

20 likes 2mo
FC
f.chowdhuryTL2 Moderator4 Jul 2026#16

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

27 likes 24d
RS
r.serranoTL2 Moderator12 Jul 2026 · edited#19

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

21 likes 16d
AM
a.mwangiTL2 Moderator17 Jul 2026#21
z.vogel, post #6: I read post #4 twice before replying, because I had assumed the opposite. What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

post #20 is right about the mechanism and I think understates the practical bit.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

26 likes in reply to #6 11d

Read the full topic (25 posts)

Suggested topics

TopicParticipantsRepliesViewsActivity
High-resolution MS on aggregates and what it can see
Posting this under the heading it deserves: High-resolution MS on aggregates and what it can see Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked.…
CRGISCBPEC+117 123 6.5k 8mo
Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. retatrutide has a…
JPERBEBWBJ+124 146 15k 16mo
Second pass at: MALDI versus ESI for a quick identity check
Second pass at: MALDI versus ESI for a quick identity check Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it…
MALKO 2 18k 7mo
Charge states for a 4 kDa peptide, worked through — the long version
Charge states for a 4 kDa peptide, worked through — the long version Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like…
HEGTMI 2 293 3mo
[2026 update] Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
GTAPFP 2 1k 8mo

Related topics — sharing the tags counter-ion content, mass error (ppm), charge states

TopicParticipantsRepliesViewsActivity
Racemisation and why it is invisible to a mass measurement
Posting this under the heading it deserves: Racemisation and why it is invisible to a mass measurement Everything below is what sits behind that. Working through the identity arithmetic and I would like it…
NTKCTDOPA+120 127 41k 1mo
High-resolution MS on aggregates and what it can see
Posting this under the heading it deserves: High-resolution MS on aggregates and what it can see Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked.…
CRGISCBPEC+117 123 6.5k 8mo
Coming back to: Oxidation pathways for methionine and tryptophan
Oxidation pathways for methionine and tryptophan Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it checked.…
YRFHKRZSCP+1 5 36k 15mo
Distinguishing a deletion sequence by mass alone
Distinguishing a deletion sequence by mass alone Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it checked.…
RMLSMHMINM+121 129 25k 17mo
MALDI versus ESI for a quick identity check
On the subject in the title: MALDI versus ESI for a quick identity check Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. tirzepatide has a…
RMTVO 2 2k 11mo