Distinguishing a deletion sequence by mass alone

This is a generated summary. It shows the 9 most-liked posts from a topic of 130, in their original order, with the accepted answer included where one exists. It is a reading aid and it will miss nuance — the full topic is the record.
NM
n.moreauTL2 Moderator Solution20 Jan 2025#5

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

10 likes 18mo
SH
s.hartmannTL2 Moderator26 Jan 2025#19
ms_holloway, post #3: Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

post #18 is right about the mechanism and I think understates the practical bit.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

30 likes in reply to #3 18mo
BB
b.brandtTL2 Moderator27 Jan 2025#23

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

32 likes 18mo
CL
customs_ledgerTL3Regular1 Feb 2025#40
b.brandt, post #23: Resolution: "high resolution" commonly means Go to post

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

33 likes in reply to #23 18mo
DS
d.szymanskiTL3Wiki editor6 Feb 2025#57

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

31 likes 18mo
EV
e.vargaTL2 Moderator7 Feb 2025#60
i.wojcik, post #37: post #36 is right about the mechanism and I think understates the practical bit. Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very… Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

30 likes in reply to #37 18mo
OF
outline_firstTL3Wiki editor9 Feb 2025#66

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

31 likes 18mo
G
GDashwoodTL3Regular12 Feb 2025#78

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

33 likes 17mo
EM
e.mensaTL2 Moderator17 Feb 2025#100
s.hartmann, post #19: post #18 is right about the mechanism and I think understates the practical bit. Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

32 likes in reply to #19 17mo

Read the full topic (130 posts)

Suggested topics

TopicParticipantsRepliesViewsActivity
[2026 update] Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. semaglutide has a…
DBBIATNAL+116 135 31k 3mo
Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. retatrutide has a…
JPERBEBWBJ+124 146 15k 16mo
About the Mass spectrometry category
LC-MS and high-resolution MS, charge states, mass accuracy in ppm, and what identity confirmation does and does not establish. This post is a community wiki: any member at trust level 3 or above can edit it,…
DOOBLDBDRV+5 9 19k 6h
Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer I have a specific reason for asking rather than idle curiosity, and the context is below. Posting the method first, because I know what the first three…
DAAPINOSS+75 82 4k 17mo
Adducts: sodium, potassium, and the peak you did not expect — does this still hold?
Asking directly, because I could not find a straight answer: Adducts: sodium, potassium, and the peak you did not expect — does this still hold? Posting the method first, because I know what the first three…
KBZYLVCETV+113 119 1.7k 3mo

Related topics — sharing the tags mass error (ppm), high-resolution MS, LC-MS

TopicParticipantsRepliesViewsActivity
Mass error in ppm: the arithmetic and the caveat
Mass error in ppm: the arithmetic and the caveat Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it checked.…
KCCIAMCDBV+108 125 1.5k 5mo
Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. retatrutide has a…
JPERBEBWBJ+124 146 15k 16mo
High-resolution MS on aggregates and what it can see
Posting this under the heading it deserves: High-resolution MS on aggregates and what it can see Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked.…
CRGISCBPEC+117 123 6.5k 8mo
Oxidation and deamidation mass shifts, tabulated
Posting this under the heading it deserves: Oxidation and deamidation mass shifts, tabulated Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked.…
AFJVIATNCS+67 76 21k 16mo
[2026 update] Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
GTAPFP 2 1k 8mo