ThibodeauTL3Regular9 Oct 2025#31

Coming back to post #29, because the follow-up matters more than the original answer.

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

16 likes 10mo
FC
f.chowdhuryTL2 Moderator14 Oct 2025#32

Picking up post #29: that is the part I would want checked first.

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

6 likes 9mo
CI
c.inglethorpeTL3Regular18 Oct 2025 · edited#33
e.ferreira, post #10: Coming back to post #8, because the follow-up matters more than the original answer. Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is… Go to post

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

1 like in reply to #10 9mo
RM
r.molnarTL2 Moderator22 Oct 2025#34

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

0 likes 9mo
C
CSagredoTL3Regular26 Oct 2025#35

I read post #33 twice before replying, because I had assumed the opposite.

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

11 likes 9mo
HR
h.ramosTL2 Moderator31 Oct 2025#36
Leiterman, post #23: This follows post #20 rather than contradicting it. Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount. Go to post

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

3 likes in reply to #23 9mo
F
FFaulknerTL3Regular4 Nov 2025#37

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

0 likes 9mo
VO
v.okonkwoTL2 Moderator8 Nov 2025#38

post #37 is right about the mechanism and I think understates the practical bit.

Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99.

32 likes 9mo
T
TamburelloTL2Member12 Nov 2025#39

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

31 likes 8mo
LL
l.lundgrenTL2 Moderator16 Nov 2025 · edited#40
revision_history, post #29: Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance. Go to post

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

16 likes in reply to #29 8mo
FN
f.novakTL2 Moderator20 Nov 2025#41
r.molnar, post #34: System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5). Go to post

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

10 likes in reply to #34 8mo
K
KTurkingtonTL3Regular24 Nov 2025#42
dr_bhattacharya, post #3: Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount. Go to post

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

22 likes in reply to #3 8mo
AN
a.norgaardTL2 Moderator28 Nov 2025#43

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

0 likes 8mo
TI
trough_indexTL3Regular2 Dec 2025 · edited#44

Worth separating two things that post #40 runs together.

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

3 likes 8mo
NL
ne.laurentTL26 Dec 2025#45
BR
buffer_reviewTL3Regular10 Dec 2025#46

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

30 likes 8mo
MA
mi.amankwahTL2 Moderator14 Dec 2025#47

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

1 like 7mo
L
LJankowiakTL3Regular18 Dec 2025#48

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

6 likes 7mo
CB
c.bakkerTL2 Moderator22 Dec 2025#49

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

21 likes 7mo
MB
m.brobergTL2 Moderator26 Dec 2025#50
BBramley, post #14: Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal). Go to post

I read post #48 twice before replying, because I had assumed the opposite.

Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99.

0 likes in reply to #14 7mo
PM
physio_marchettiTL2Physiotherapist30 Dec 2025#51

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

1 like 7mo
NO
n.oseiTL2 Moderator3 Jan 2026#52

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

0 likes 7mo
CL
coldchain_liuTL3Regular7 Jan 2026#53

I read post #51 twice before replying, because I had assumed the opposite.

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

18 likes 7mo
JI
j.iyerTL2 Moderator11 Jan 2026#54
m.broberg, post #50: I read post #48 twice before replying, because I had assumed the opposite. Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99. Go to post

This follows post #51 rather than contradicting it.

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

7 likes in reply to #50 7mo
TY
two_year_lineTL3Regular14 Jan 2026 · edited#55
ne.laurent, post #45: Picking up post #42: that is the part I would want checked first. Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

0 likes in reply to #45 6mo
SK
s.kuuselaTL2 Moderator18 Jan 2026#56

post #55 answers the question as asked. The question underneath it is different.

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

0 likes 6mo
MM
maintenance_modeTL3Regular22 Jan 2026#57

Coming back to post #55, because the follow-up matters more than the original answer.

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

12 likes 6mo
AP
au.pereiraTL2 Moderator26 Jan 2026#58

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

4 likes 6mo
LW
l.wikstromTL2 Moderator29 Jan 2026#59

Worth separating two things that post #55 runs together.

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

0 likes 6mo
This topic was closed 120 days after the last reply. Closing is automatic for quiet topics so that a settled answer does not collect new questions underneath it. If you have a follow-up, open a new topic and link back to this one — that keeps both readable and gives your question its own title.

Suggested topics

TopicParticipantsRepliesViewsActivity
An inter-laboratory comparison on one lot, tabulated
On the subject in the title: An inter-laboratory comparison on one lot, tabulated Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. semaglutide has a…
ZAMGCRKB 3 17k 10mo
About the Method validation category
Specificity, linearity, precision, accuracy, robustness, system suitability, and why two labs disagree. This post is a community wiki: any member at trust level 3 or above can edit it, and every edit is…
KTVPAKAJV+1 5 468 5mo
Revisiting: Sample handling between despatch and injection
Posting this under the heading it deserves: Revisiting: Sample handling between despatch and injection Everything below is what sits behind that. Posting the method first, because I know what the first three…
DVEAGTEMNG+5 9 1.1k 6mo
Second pass at: LOD and LOQ: definitions people get wrong
Posting this under the heading it deserves: Second pass at: LOD and LOQ: definitions people get wrong Everything below is what sits behind that. Working through the identity arithmetic and I would like it…
SPNIBIMG+47 51 40k 10mo
When to suspect the method rather than the sample
When to suspect the method rather than the sample — that is the question, and I have not found it answered plainly anywhere I have looked. Posting the method first, because I know what the first three replies…
MMCCYMRMBJ+109 115 63k 15mo

Related topics — sharing the tags linearity, LOD & LOQ, recovery

TopicParticipantsRepliesViewsActivity
When to suspect the method rather than the sample
When to suspect the method rather than the sample — that is the question, and I have not found it answered plainly anywhere I have looked. Posting the method first, because I know what the first three replies…
MMCCYMRMBJ+109 115 63k 15mo
Reflectance and photometric devices: precision in practice
On the subject in the title: Reflectance and photometric devices: precision in practice Working notes rather than a conclusion. I would like to understand what this number means before I repeat it anywhere. A…
RMGVB 2 2.2k 10mo
About the Method validation category
Specificity, linearity, precision, accuracy, robustness, system suitability, and why two labs disagree. This post is a community wiki: any member at trust level 3 or above can edit it, and every edit is…
KTVPAKAJV+1 5 468 5mo
Reading a home test result without over-claiming — what changed since
On the subject in the title: Reading a home test result without over-claiming — what changed since Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked.…
KDPMHV 2 12k 6mo
Coming back to: What consumer-accessible tests can and cannot determine
Asking directly, because I could not find a straight answer: What consumer-accessible tests can and cannot determine Posting the method first, because I know what the first three replies will otherwise be.…
PKNWVNHK+4 8 25k 22mo